RotemNet software ver 4.20 installed on PC i5 windows 7 pro 4 gb ram . Usb connection to Communicator
ver 3.04 hardware ver 3.00 . Connected to Rlink Rs 485 at communicator to Rlink Rs 232 at controllers ( Percision ver 3.06 ver. ) Phase 1 is 4 controls . At initial scan all 4 controls scanned ok with data downloaded quickly . Accessing all menus ok within seconds , after 5 min operating ok then comm error on all 4 controls . Will not connect again until PC is rebooted or USB unplugged and reconnected then communicate great for 5 min then same comm error . Also may crash PC ( Blue screen ) Test menu on communicator shows good comm with all controls .

0
RotemTech

Moderator
Registered:
Posts: 168
Reply with quote  #2 
4.02.03 Is the latest software so I'd get that if you don't already have it.  
Make sure you also have their latest USB driver also.
Make sue you are running the program as administrator (right click properties).
From there it's on to PC settings making sure your USB ports aren't going in to a sort of  a low power standby mode etc.  Making sure there is an allowance to allow RotemNet through the firewall in windows, and also making sure there are no virus protection programs such as McAfee that may be interfering.
Give us a call at Diversified Imports if you've exhausted all other resources.  Something else to consider is maybe purchasing a serial to USB adapter and using that instead to see if that helps.  Also keep in mind RotemNet should be exited (using the door icon) when not in use.  It's not really meant to run all the time.
0
Uastech

Registered:
Posts: 2
Reply with quote  #3 
This Communication issue has been solved by using Usb to serial adapter and drivers supplied with adapter . Thanks
0
RotemTech

Moderator
Registered:
Posts: 168
Reply with quote  #4 
Nice, I'm glad to hear it.  The serial port is a solid near bullet proof solution if the USB goes out.
0
Marion

Registered:
Posts: 10
Reply with quote  #5 
Hi Diversified Team,


Just like to quote your feedback: Also keep in mind RotemNet should be exited (using the door icon) when not in use.  It's not really meant to run all the time.


^^If that is the case, are you saying that using SAVE MAINSCREEN DATA EVERY 1 MIN OR SO, is bad? Not advisable?

Regards,
Marion San Pedro
0
RotemTech

Moderator
Registered:
Posts: 168
Reply with quote  #6 
Marion, what you are doing sounds fine.  For best results in your scenario make sure the PC is dedicated for RotemNet use.
0
Marion

Registered:
Posts: 10
Reply with quote  #7 
Noted with Thanks!

Marion